To reduce patient dose, exposure technique in digital imaging should be adjusted by which change?

Prepare for the Mosby Digital Image Acquisition Test with confidence. Utilize comprehensive flashcards and multiple-choice questions, each equipped with hints and explanations. Ensure your readiness for the exam!

Multiple Choice

To reduce patient dose, exposure technique in digital imaging should be adjusted by which change?

Explanation:
Increasing kVp reduces patient dose by making the x-ray beam more penetrating, so you can lower the number of photons (mAs) needed to achieve the same detector exposure. In digital imaging, detectors handle lower contrast and still provide diagnostic quality, so you can raise kVp without sacrificing image usefulness while cutting dose. Lowering kVp would require more mAs to reach the detector, boosting dose; increasing mAs directly raises dose; shortening SID increases dose due to the inverse square law and usually necessitates more exposure to maintain receptor exposure. Therefore, raising kVp is the effective way to reduce patient dose.

Increasing kVp reduces patient dose by making the x-ray beam more penetrating, so you can lower the number of photons (mAs) needed to achieve the same detector exposure. In digital imaging, detectors handle lower contrast and still provide diagnostic quality, so you can raise kVp without sacrificing image usefulness while cutting dose. Lowering kVp would require more mAs to reach the detector, boosting dose; increasing mAs directly raises dose; shortening SID increases dose due to the inverse square law and usually necessitates more exposure to maintain receptor exposure. Therefore, raising kVp is the effective way to reduce patient dose.

Subscribe

Get the latest from Examzify

You can unsubscribe at any time. Read our privacy policy